We recently purchased several new serologic cell washers. During validation testing, we have found that the cell button is not as "dry" as on our previous equipment from the same manufacturer (IEC). We are on our 3rd replacement of the new washer. The first one received had several drops of saline left on the cells which was absolutely unacceptable. This last replacement leaves a very small amount of saline on the red cell button, but is easily seen as different than our other washers. What is a good guideline for acceptability or definition of a good decant on the cell button?

  • Replies 1
  • Views 3.4k
  • Created
  • Last Reply

Top Posters In This Topic

comment_9383

The purpose of the cell washer is to automatically remove protein from the cells, not to produce a dry button. The automated decant cycle of all cell washers I've used will inevitably leave a small amount of saline, as will a manual decant unless you're really, really good at blotting it. I guess the question is how much is too much, and we defined it as one drop (25uL?)

A small amount of residual saline will not affect the test results. DAT testing can be performed with a 3% cell suspension, which is almost all saline in the drop you add to the tube.

Since moving to gel, we hardly ever use our cell washer anymore ...

Create an account or sign in to comment

Recently Browsing 0

  • No registered users viewing this page.

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.

Configure browser push notifications

Chrome (Android)
  1. Tap the lock icon next to the address bar.
  2. Tap Permissions → Notifications.
  3. Adjust your preference.
Chrome (Desktop)
  1. Click the padlock icon in the address bar.
  2. Select Site settings.
  3. Find Notifications and adjust your preference.