Jump to content

Featured Replies

Posted
comment_8969

We are currently standarizing dig therm monthly and would like to move to quarterly or even semi-annually. Who has experience with this type of variance from the FDA? What data did you use to support the request? etc. Thanks for any help.

  • Replies 4
  • Views 4k
  • Created
  • Last Reply

Top Posters In This Topic

comment_8993

I check the calibration on my digital thermometers yearly, not monthly or even quarterly. I assume you are referring to the table in 21 CFR 606.60 which defines the frequency of calibration for digital thermometers as monthly. Part of the problem with the CFR is that it hasn't been updated in quite a while and time and technology march on. The manufacture usually has a recommend frequency of calibration with devices of this type and while digital inputs (once upon a time) experienced a lot of drift, such is no longer the case. We did not file for a variance for 21 CFR 606.60, we just made certain our documentation is in place. Our most Recent FDA inspection was in June of this year, our inspector did look at our calibration records for thermometers and refrigerators and we did not receive a 483. Typically, if you can document and validate your process, proving it is in control, there is no issue.

Of course, we are a registered (not licensed) blood bank and transfusion service. That could have an impact as the FDA looks at licensed manufactures a little differently. So your mileage may vary.

comment_9019

We check ours annually also. If you have yours calibrated by Control Company, they will put a re-calibration required in one year label on the thermometer. I send ours off to get it back in time to re-calibrate all our other thermometers in December.

comment_9046

I suspect that tgriffin is referring to the digital thermometer calibration checks required for thermometers used to qualify donors. The regs do require monthly calibration checks for those thermometers.

As the FDA what they require for a variance waiver. That is your best source. That said, I don't know of any donor center that has been given a waiver.

Good luck!

BC

comment_9058

If referring to 21 CFR 606.60(B) a variance request is no longer needed when following manufacturer's instructions for a digital thermometer. See the link below:

http://www.fda.gov/cber/blood/exceptions.htm

Hope this helps.

Create an account or sign in to comment

Recently Browsing 0

  • No registered users viewing this page.

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.

Configure browser push notifications

Chrome (Android)
  1. Tap the lock icon next to the address bar.
  2. Tap Permissions → Notifications.
  3. Adjust your preference.
Chrome (Desktop)
  1. Click the padlock icon in the address bar.
  2. Select Site settings.
  3. Find Notifications and adjust your preference.