Can immediate spin be eliminated for cold agglutinins? We are going back and forth with this because immediate spin detects ABO incompatibility but the computer system is also set up to detect this.

  • Replies 4
  • Views 693
  • Created
  • Last Reply

Top Posters In This Topic

Most Popular Posts

  • SbbPerson
    SbbPerson

    Are you doing the immediate spin for a crossmatch or antibody screen? Well, we use the prewarm method for Cold agglutinins. We enter the results to the prewarm technique which is Negative, so the comp

  • Bet'naSBB
    Bet'naSBB

    At our facility, we have to do a cold adsorption IF the cold auto interferes with the reverse and/or ISXM. Since testing with "neat" plasma is our standard of practice, we would still report the XM as

  • AuntiS
    AuntiS

    In my opinion, if you have: A cold antibody (not pathological), and Have done the workup to show that there is no other antibody present (i.e. a clinically significant antibody such as anti-E), and

comment_94605

Are you doing the immediate spin for a crossmatch or antibody screen? Well, we use the prewarm method for Cold agglutinins. We enter the results to the prewarm technique which is Negative, so the computer doesn't have a cow. Then we write in the test comments that the prewarm technique was used due to cold agglutinin, or history of cold agglutinin.

Edited by SbbPerson

  • Author
comment_94641

The immediate spin is for crossmatch. Thank you for your input. We are currently doing the prewarm method for colds also but our QA is concerned about prewarming so there are discussions to eliminate the I.S. phase.

comment_94658

At our facility, we have to do a cold adsorption IF the cold auto interferes with the reverse and/or ISXM. Since testing with "neat" plasma is our standard of practice, we would still report the XM as incompatible and send it out with a release stating the incompatibility was due to an autoantibody. We are not "allowed" to prewarm away IS reactions. 🥶 IF we prewarm - it's only if the cold has a high thermal amplitude and causes interference after 37deg incubation / antiglobulin phase...........but that's just us.........

comment_94664

In my opinion, if you have:

  • A cold antibody (not pathological), and

  • Have done the workup to show that there is no other antibody present (i.e. a clinically significant antibody such as anti-E), and

  • Your LIS dictionary set up so that your reported antibody is not deemed as clinically significant

You can set it up so that the computer (electronic) crossmatch can be used.

If it is a pathological antibody, we use incompatible (i.e. Least incompatible in Meditech)

Create an account or sign in to comment

Recently Browsing 0

  • No registered users viewing this page.

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.

Configure browser push notifications

Chrome (Android)
  1. Tap the lock icon next to the address bar.
  2. Tap Permissions → Notifications.
  3. Adjust your preference.
Chrome (Desktop)
  1. Click the padlock icon in the address bar.
  2. Select Site settings.
  3. Find Notifications and adjust your preference.