We switched to N-Hance years ago in order to comply with Massachusetts Water Resources regulations. ImmuAdd contains the preservative thimerosol-a mercury contaminant. We performed a small validation and found that N-Hance performed as well as ImmuAdd. Only one problem: the manufacturer's insert suggests using a transfer pipet to dispense the additive rather than the vial dropper in order to maintain the proper ionic strength. When we open a vial, we always discard the dropper and replace it with a stopper.

  • Replies 30
  • Views 7.5k
  • Created
  • Last Reply

Top Posters In This Topic

Most Popular Posts

  • Malcolm Needs
    Malcolm Needs

    When I first joined the wonderful world of blood transfusion, with particular reference to blood group serology, at the International Blood Group Reference Laboratory, when it was in London, my mentor

  • Marilyn Plett
    Marilyn Plett

    I don't believe that an optical aid necessarily refers to a microscope. In my pre-retirement years we used the agglutination viewer when an optical aid was required. Example: https://www.fishersci.com

  • David Saikin
    David Saikin

    As a neophyte blood banker we read everything under the scope (early 70s - it was SOP).  Then took a job and worked w a very experienced blood banker (she was on a first name basis w Laurie Marsh, the

Posted Images

comment_80815
2 hours ago, Monique said:

We switched to N-Hance years ago in order to comply with Massachusetts Water Resources regulations. ImmuAdd contains the preservative thimerosol-a mercury contaminant. We performed a small validation and found that N-Hance performed as well as ImmuAdd. Only one problem: the manufacturer's insert suggests using a transfer pipet to dispense the additive rather than the vial dropper in order to maintain the proper ionic strength. When we open a vial, we always discard the dropper and replace it with a stopper.

For reference:

image.png.1d095d74848ca418e949736e1269078e.png

Edited by AMcCord

comment_80824
2 hours ago, AMcCord said:

For reference:

image.png.1d095d74848ca418e949736e1269078e.png

Wow. I perfectly understand the science, but that is an awful thing to put in a Directions for Use. A savvy Inspector could throw serious doubt on any tests performed using the supplied dropper. And why provide a dropper if it isn't good enough for the test ? The only way to meet this requirement to the fullest is to use a calibrated semiautomatic pipette.

comment_80838
21 hours ago, exlimey said:

Wow. I perfectly understand the science, but that is an awful thing to put in a Directions for Use. A savvy Inspector could throw serious doubt on any tests performed using the supplied dropper. And why provide a dropper if it isn't good enough for the test ? The only way to meet this requirement to the fullest is to use a calibrated semiautomatic pipette.

Or is the question the quality of the disposable pipette that is used for the patient plasma? We do use a disposable intended for Blood Bank use with a statement on the box  'consistent drop size'. I would say that you could present documentation that you've verified volume delivered (on average - since its a disposable) by the pipette vs the dropper in the vial.

comment_80845
22 hours ago, AMcCord said:

Or is the question the quality of the disposable pipette that is used for the patient plasma? We do use a disposable intended for Blood Bank use with a statement on the box  'consistent drop size'. I would say that you could present documentation that you've verified volume delivered (on average - since its a disposable) by the pipette vs the dropper in the vial.

That's an excellent point. It always struck me as slightly odd that such critical testing is done by "drops" - a potentially highly variable volume. One certainly wouldn't see an HIV test kit give instructions like "Dispense two drops". Goes to show that the standard serological (tube) test is extremely tolerant of variation.

comment_80895
On 8/21/2020 at 6:07 AM, exlimey said:

That's an excellent point. It always struck me as slightly odd that such critical testing is done by "drops" - a potentially highly variable volume. One certainly wouldn't see an HIV test kit give instructions like "Dispense two drops". Goes to show that the standard serological (tube) test is extremely tolerant of variation.

True, but I suppose we can also say that there has been years and years worth of testing performed that way with no evidence of harm reported from the practice.

Create an account or sign in to comment

Recently Browsing 0

  • No registered users viewing this page.

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.

Configure browser push notifications

Chrome (Android)
  1. Tap the lock icon next to the address bar.
  2. Tap Permissions → Notifications.
  3. Adjust your preference.
Chrome (Desktop)
  1. Click the padlock icon in the address bar.
  2. Select Site settings.
  3. Find Notifications and adjust your preference.