comment_78464

I'm in the process of validating our new Ortho vision Analyzer and have just run our latest CAP specimens however I'm getting false positives on the back groups. Does anyone else use this analyser? are there any know issues with CAP specimens?

 

Thanks 

  • Replies 10
  • Views 3.2k
  • Created
  • Last Reply

Top Posters In This Topic

Most Popular Posts

  • JAT-C  Got to the bottom of it with one of them, it was Anti-c antibody positive and was cold reactive with the A cells. Switched for some c- cells and now clear.

Posted Images

  • Author
comment_78507

JAT-C 

Got to the bottom of it with one of them, it was Anti-c antibody positive and was cold reactive with the A cells. Switched for some c- cells and now clear.

comment_78516

Thats unusual, the anti small c in that sample should only be reactive at IAT phase.  You could see if it is reactive in immediate spin with rr cells and if so advise CAP about it.

  • Author
comment_78547
6 hours ago, noelrbrown said:

Thats unusual, the anti small c in that sample should only be reactive at IAT phase.  You could see if it is reactive in immediate spin with rr cells and if so advise CAP about it.

I carnt really see any other explanation for it and would be interested to hear from any other Ortho users who have this latest CAP.  Yes the backgroup was clear on Diamed/Bio-rad and they too had a c Poitive A cell,  I tried the same cells on Ortho and it was Poitive, then switched to a screened A cell that was c Neg and it was clear. I didnt panel the antibody but going off the screening cell result it can only be anti-c.

comment_78559

The Ortho cards use a potentiator for the back type section.  This potentiator may be causing the phenomenon.

comment_78611

It is against CAP regulations to be discussing survey results before the due date.

comment_78614

Why aren't you using the CAP Survey specifically designed for automated instruments? I believe it is CAP-J.

comment_78615
On 10/3/2019 at 1:20 PM, srichar3 said:

JAT-C 

Got to the bottom of it with one of them ...

 

2 minutes ago, Dansket said:

Why aren't you using the CAP Survey specifically designed for automated instruments? I believe it is CAP-J.

 

  • 3 weeks later...
  • Author
comment_78753
On 10/12/2019 at 6:45 AM, Dansket said:

Why aren't you using the CAP Survey specifically designed for automated instruments? I believe it is CAP-J.

 

JAT is the automated survey. I think J is a manual only survey. 

On 10/11/2019 at 8:43 PM, swede said:

It is against CAP regulations to be discussing survey results before the due date.

I wasn't really discussing simply asking if there are know issue with CAP samples on Ortho Technology, as certain technologies do have know issues with EQA specimens. As this analyser isn't in use yet we are reporting the results from the IH anyway. 

 

ScreenHunter_055.jpg

comment_78777

No one should discuss any CAP or API result before due dates. You should not discuss even you reported your results. You need to wait until CAP due date is over. This is really against CAP regulation to discuss here. 

Create an account or sign in to comment

Recently Browsing 0

  • No registered users viewing this page.

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.

Configure browser push notifications

Chrome (Android)
  1. Tap the lock icon next to the address bar.
  2. Tap Permissions → Notifications.
  3. Adjust your preference.
Chrome (Desktop)
  1. Click the padlock icon in the address bar.
  2. Select Site settings.
  3. Find Notifications and adjust your preference.