    mminhas44 reacted to Veejay in Gel cards   
    We did it easily by putting the gel card centrifuge in a 37C dry air incubator along with all the consumables. We waited 30 minutes to get all up to temp then did the testing. Worked great.
     
  2. Sad
    mminhas44 reacted to AMcCord in Gel cards   
    I tried it several times when we were still gel users, but never successfully.
  3. Like
    mminhas44 reacted to Malcolm Needs in Gel cards   
    It can, but it is exceptionally difficult to get the air in the reaction well up to a strict 37oC, and "cold antibodies" sensitise red cells VERY quickly.
  4. Like
    Sorry - I forgot the C.
    As the frequency of C is about 68%, you can remove 2 of those three.  That leaves you with 1 unit
    Anna
  5. Like
    Very very roughly, and assuming your donors are mostly White European
    If you start off with 100 units, approx. 50 will be O or B
    Of those approx. 5 will be K+ - so you will have 45 units.
    Of those, about 2/3  (so approx. 30) will be Fya+, so you will have 15 left
    Of those about 80%  (about 12) will be Jka+
    So, if you're VERY lucky, the remaining 3 will be compatible.
    Good luck!!!!!
  6. Like
    mminhas44 reacted to Hajar in High Performance Liquid Chromatography (HPLC)   
    U need to exclude recent transfusion or carry over from previous sample ( although the levels mentioned is high for carry over)
    wise if above mentioned are excluded to consider genetic testing and parental  screening 
  7. Like
    mminhas44 got a reaction from Baby Banker in Visual Inspection   
    Could be platelet clumps..
  8. Like
    mminhas44 reacted to kimannez in Automated CSF counts   
    cells/microliter
     
  9. Like
    mminhas44 reacted to kimannez in Automated CSF counts   
    Could you be more specific? Are you asking about units, or whether results need to be verified?
     
  10. Like
    mminhas44 reacted to kackieanne in malaria detection by automated analyser....?   
    I saw a Blood Hound prototype 5-10 years ago.  I wish the FDA would work faster.  I think the technology looks very very hopeful. 
  11. Like
    mminhas44 reacted to Joseph in malaria detection by automated analyser....?   
    I just saw the Blood Hound at the AACC. It does digital images of all the WBC's, RBC's and Plts on a CBC. Per the manufacturer, it is capabel of detecting and counting RBC inclusions, including malaria, in RBC's. It looked impressive and is pending FDA approval. Check out their website.
  12. Like
    mminhas44 reacted to Townsend in Subgroup in a Neonate?   
    Agree - also, what blood type is the mother?  From an operational stand-point you have a couple of options: Call the blood type "indeterminate" or "unable to determine" or call the baby AB.  Either way, we would suggest repeat testing at 4-6 months and include A1 lectin typing if discrepancy still exists.  In the meantime we would give group O red cells and AB plasma/platelets until resolved.
  13. Thanks
    mminhas44 reacted to Alan Neal in QC   
    Agree - highly probably that the SD being used is not appropriate i.e. to large. 
    Results should be distributed across the L-J chart, with approx. 1:20 results in the 2-3 SD space.
    Suggest review local and published SD for analytes
  14. Thanks
    mminhas44 reacted to SMILLER in QC   
    It sounds like your SD is set too low.  Try recalculating it based on recent results.
    Scott
  15. Like
    mminhas44 reacted to carol1 in Chronic Reactive Lymphocytosis   
    I'm not an expert, but flow cytometry may be helpful to classify the lymphocytes (B-cell, T-cell, NK cells) and see if an abnormal immunophenotype exists.
  16. Thanks
    mminhas44 reacted to SMILLER in Verification of Vesmatic ESR analyzer   
    I am not familiar with ISO 15189 standards--at a minimum you obviously have to follow them, along with manufacturer's requirements--but it seems to me...
    ...like you are being advised to validate a particular analyzer by comparing it to another exactly the same, which presumably was compared to another exactly the same, which was...
    With any new method, you must show correlation with a previous method, in this case, your manual method.  Comparing it to an analyzer at another facility seems rather pointless.  (The exception being if the other facility is part of your system, and you use the same procedures and reference ranges).
    You should also be doing a reference range study to validate your reference range, especially if there is significant variance in the correlation.  It has been our experience that automated methods correlate pretty well with the old manual ESRs.
    Scott

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.

Configure browser push notifications

Chrome (Android)
  1. Tap the lock icon next to the address bar.
  2. Tap Permissions → Notifications.
  3. Adjust your preference.
Chrome (Desktop)
  1. Click the padlock icon in the address bar.
  2. Select Site settings.
  3. Find Notifications and adjust your preference.