We've been live with Cerner since February 2013 and our process is the same. We get a unit of the patient's blood type from the fridge, go into Result Entry, scan the sample accession label, scan the barcoded BBID (blood bank i.d. from the armband) on the sample, and then scan the unit (however many are ordered). As soon as you scan a unit, the computer makes the decision that the unit is ABO compatible with the recipient and automatically fills in the interpretation field with Computer XM OK. Don' forget to pull a segment from the unit as well as a number sticker--we affix the sticker on a 12X75 tube and place the segment inside folded so it won't fall out. Often we will wait until a nurse comes to pick up blood and just perform a Computer Crossmatch Dispense at that time. The only thing is you have to manually enter the BBID in the field in the save dialog box when you are dispensing so that it will print on the donor tag. Since we do not have the sample in hand at that time, I had to figure out a way to be able to quickly find the blood bank armband i.d. What I did was to build a BBID result field in the ABORh test so that we can quickly look in Order Result Viewer at the most recent type and screen results and see the BBID. We can then verify it against the blood request form that the nurse brings which MUST have the BBID on it. We type this into the BBID field during dispense, pull a segment and sticker and we're done. We never have to handle the patient's sample again. It took us awhile to get used to this as we went live with electronic crossmatching the same time we went live with Cerner. Everyone loves it now and we've never looked back, including this "old timer" blood banker!