  1. Like
    TRabs10 reacted to gene20354 in GEL DAT CONTROL   
    We run the monoclonal control card in parallel with the poly card to catch the false positives.
  2. Like
    TRabs10 got a reaction from bldbnkr in IFU Anti-D   
    One of our sister sites was recently cited by HFAP for not following their Anti-D package insert (BioRad Seraclone Anti-D blend for tube testing). Their procedure is to interpret any obstetrical RH type that reacts 1+ or less at immediate spin NEGATIVE until they refer the sample for genotyping. A chartable comment is added to the result stating that the reaction was discordant and that the sample was referred for genotyping.  If genotyping shows a Partial D, the interpretation stays as Negative. If the genotyping comes back as weak D 1, 2, 3, or D+ the interpretation is corrected to Positive and a corrected report sent. However, the BioRad package insert says "agglutination of red cells is a positive result". hence the citation. We realized that the "less than 2+" rule in the procedure refers to the FORMER Anti-D reagent (Immucor Series 4) and the SOP was left that way when they switched to BioRad. They aren't interested in changing the current process, because it has helped identify at least 3 partial D mothers who would not have received RhIG if the insert had been followed to a T. My question is, does anyone else follow a similar "less than 2+" rule, have you been cited, and what Anti-D reagent are you using? Do you result these patients at "Indeterminate", at least until genotyping is complete?
     
    Thank you!
  3. Like
    TRabs10 got a reaction from Malcolm Needs in Rh Pos or Rh Neg?   
    You may be interested in "Strategies to identify candidates for D variant genotyping" by Luo, Keller, et. al. (Blood Transfus 2018; 16: 293-301). Our facility currently uses BioRad Anti-D via tube testing. If an OB patient reacts 1+ or less at immediate spin we send the patient out for Weak D molecular testing. We just got an Ortho Vision, so I am looking at implementing the strategies outlined in the paper to help identify those Weak or Partial D patients.

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.

Configure browser push notifications

Chrome (Android)
  1. Tap the lock icon next to the address bar.
  2. Tap Permissions → Notifications.
  3. Adjust your preference.
Chrome (Desktop)
  1. Click the padlock icon in the address bar.
  2. Select Site settings.
  3. Find Notifications and adjust your preference.