  1. COM.10300 Knowledge of Policies and Procedures Phase II The laboratory has a defined process and records indicating that all personnel are knowledgeable about the contents of the policies and procedures (including changes) relevant to the scope of their testing activities. NOTE: The form of this system is at the discretion of the laboratory director. Annual procedure sign-off by testing personnel is not specifically required. Evidence of Compliance: ✓ Records indicating that the testing personnel have read the policies and procedures, new and revised, OR records of another written method approved by the laboratory director
  2. Ortho Vision Not Reading Barcodes

    BB Gal reacted to BldBnker for a post in a topic

    1 point
    We had issues with our Visions not reading barcodes also. We did learn that the Vision needs approximately 4 mm of a white margin on either side of the barcode to read. We adjusted our barcode printer and that enabled the Visions to read the label. Not sure if anyone at Ortho told you that. I think it was our service rep that let us know that hint! Good luck!
  3. Ortho Vision Not Reading Barcodes

    Bet'naSBB reacted to Cliff for a post in a topic

    1 point
    It's been a few years, but the 800+ bed hospital I was at had two IH1000. I'm not saying they were perfect, but they were workhorses, and I don't remember them not reading or misreading the barcodes.
  4. Ortho Vision Not Reading Barcodes

    RRay reacted to John C. Staley for a post in a topic

    1 point
    If I were you I would cut my loses and go with a different company for my next analyzer. I never did like anything about Ortho and avoided dealing with them any time I had the chance. Just my opinion for what ever it's worth.
  5. 3% to 0.8% conversion

    SbbPerson reacted to RRay for a post in a topic

    1 point
    I have the official conversion guide from Ortho and it doesn't recommend any specific QC apart from the general QC of the method it alludes to (the general MTS procedure manual). I mention visual inspection of cell button size in my SOP. Has always covered it in the past, but you know how it goes... Quick Reference Quide - Conversion.pdf
  6. Ortho Vision Not Reading Barcodes

    BldBnker reacted to RRay for a post in a topic

    0 points
    @BldBnker Yes, Ortho did tell us that and we have more than 4mm on each side, closer to 8-10mm. Even some barcodes that don't have the white space, say because of date/initials in the way or blood smear, will often read when the perfectly printed and clean labels will not. It's making me crazy.

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.

Configure browser push notifications

Chrome (Android)
  1. Tap the lock icon next to the address bar.
  2. Tap Permissions → Notifications.
  3. Adjust your preference.
Chrome (Desktop)
  1. Click the padlock icon in the address bar.
  2. Select Site settings.
  3. Find Notifications and adjust your preference.