We currently use a procedural control for our manual cell counts.  We are currently CAP certified, but are switching to Joint Commission.  One of their Standards states that for manual cell counts, one level of QC for every 8 hours of testing is performed and documented.  Does anyone accredited by Joint commission use procedural controls?

Share this post


Link to post
Share on other sites


We do indeed run wet controls when we do cell counts on the hemacytometer.  We use Streck Cell-chex.  JCAHO noted that we needed to do this a few inspections ago.  Once every 8 hours when testing patients.

(Not sure what you mean  by a procedural control vis a vis manual cell counts...)

Scott

Share this post


Link to post
Share on other sites

our procedural control is correlation of the cell count with the cellularity of a stained slide.  CAP allows this.

Share this post


Link to post
Share on other sites

Really?  So if there are cells on the slide and cells on the hemacytometer then you are good for QC?  Those CAP dudes are far-out! Crazy Man!

Scott

Share this post


Link to post
Share on other sites

Yes!  There are guidelines that correlate the cell count with the cellularity of the cytospin.

We will be switching from CAP accreditation to Joint commission.  June will be our first inspection.  This is why I am asking about the QC.

Share this post


Link to post
Share on other sites
21 hours ago, dtreadway said:

Yes!  There are guidelines that correlate the cell count with the cellularity of the cytospin.

 

I would think that it would have to be arbitrary to the point of uselessness, but what do I know?

In any event, we are not CAP certified but have been inspected by JCAHO forever.  I am pretty sure you will need cell-count QC like mentioned above.  You can always check your inspection standards when you get a copy!

Scott

Share this post


Link to post
Share on other sites

We us a body fluid control from R&D when doing a cell count on a hemacytometer. There are 2 levels in the control kit, we alternate levels when we get a fluid. The tech who is doing the count is also required to do one level of QC.

 

Lisa

Share this post


Link to post
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
Sign in to follow this  

  • Advertisement

×

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.