Timestrip Plus BT6


rdlash

Recommended Posts

Hello all,

I have seen a few posts mentioning the Timestrip Plus BT6 and wanted to see if someone is still around that uses them.  We are switching to the BT6 from the BT10 and are having major problems with the devices triggering.  The arming procedure is different for the BT6's and we've followed the manufacturers suggestions and are still getting triggered devices that should not be triggering.  Any guidance would be greatly appreciated.

Thank you,

Ryan

Link to comment
Share on other sites

I'm still validating them. At first we would see a sliver of blue at 4-5°C, but I think we figured it out. After activation, it has to be stored at </= 3°C. If we put it on a frozen ice pack for 15 minutes, it doesn't activate at lower temps. The problem I'm having is it will start to activate at 5-5.5°C. We can temp it with an infrared gun to avoid wasting the unit, but will have to write the policy to reflect that process. 