Validation of Other Manufacturer's Reagent Red Cells for Use in Manual Grifols Gel


Quality_Lady

Recommended Posts

Has anyone validated the use of other manufacturer's (Ortho, Immucor, Bio-Rad, Quotient) reagent red cells in the Grifols gel system.  I realize that they need to be washed, and resuspended to 0.8% in Grifols diluent.  What I am looking for is general information on the actual validation performed -

  • how many cells
  • what antigen groups
  • distribution between positive and negative cells
  • use of patient samples for testing

Also - what is probably more of an issue.  Have you been inspected by FDA (or CAP, AABB) and has your plan been accepted?

Thanks in advance for your help.

 

Link to comment
Share on other sites

4 hours ago, galvania said:

Regardless of how many tests you carry out to validate, please remember one thing.  If you have problems with results in the future, neither manufacturer will take responsibility unless their IFUs specifically state that you are allowed to cross over like this

A very good point, Anna. The same applies to anything done "off-label", regardless of how much validation has been done.

Link to comment
Share on other sites

Out of curiousity, which manufacturer are you trying to validate?

Another hospital in my town validated ortho cells in the grifols gel. We played with it, but they always had a small trail of cells in the negative reactions. Nothing I was comfortable using.

I've considered validating the quotient cells in the grifols gel. I've been told they work well together.

Link to comment
Share on other sites

On ‎1‎/‎20‎/‎2018 at 6:40 PM, Carrie Easley said:

We validated Quotient  panel cells for use in Grifols DG gel.  It’s nice to get more rule-out’s/in’s while still using the same method.

Carrie - would you consider sharing your plan - or at least a high level summary of your rationale for the number and distribution of samples / cells tested?

Thanks!

Link to comment
Share on other sites

9 hours ago, Quality_Lady said:

Carrie - would you consider sharing your plan - or at least a high level summary of your rationale for the number and distribution of samples / cells tested?

Thanks!

I'd like a copy too if you don't mind :)

Link to comment
Share on other sites

4 hours ago, MaryPDX said:

We use Immucor cells in Grifols gel cards and they work fine.  (we use them for additional rule outs when necessary, not the first line of the ABID.)

Can you share your validation plan? 

Link to comment
Share on other sites

On ‎1‎/‎26‎/‎2018 at 6:59 PM, MOBB said:

Can you share your validation plan? 

  We've been doing it for over 15 years (from the time we had Ortho's gel cards).  Whatever validation was recorded, it's from eons ago and I wasn't involved with that one.

I would imagine it was similar to the validation done with the DAT method in gel cards (from my other post in that section)

Edited by MaryPDX
Link to comment
Share on other sites

On 1/28/2018 at 9:27 AM, Eagle Eye said:

We include a step in our method corelation (twice a year) where we run Ortho gel panel (0.8%), Tube panel (immucor 3%) and Immucor gel panel(dilute immucor to 0.8%). Our method was correlated almost 20 years ago!

How many antibodies do you include in your method to method?

Link to comment
Share on other sites

For method correlation: only two sample. one positive and negative. Usually postive has one of two antibodies.

Initial validation was done many years ago but if i have to do it now, i will do total 20 specimens and at least five or more positive with different antibodies (Anti-E, anti-K, ANti-D due to RHogam, warm auto and atleast one example of cold antibody)

Link to comment
Share on other sites

  • 2 months later...

For those of you who dilute 3-5% reagent RBC solution to a 0.8% solution what dilution process do you follow? what volumes of 3-5% suspension or packed RBCs and diluent do you use? The Ortho/MTS instructions for use gives the example of 10µL of packed RBCs in 1.0mL of diluent but leaves it open to other processes and preparing 1.0mL of cell suspension per reagent RBC tested seems excessive.

Link to comment
Share on other sites

19 hours ago, Lucky Jack said:

For those of you who dilute 3-5% reagent RBC solution to a 0.8% solution what dilution process do you follow? what volumes of 3-5% suspension or packed RBCs and diluent do you use? The Ortho/MTS instructions for use gives the example of 10µL of packed RBCs in 1.0mL of diluent but leaves it open to other processes and preparing 1.0mL of cell suspension per reagent RBC tested seems excessive.

I think most would agree that using up 10uL of reagent RBCs is not excessive.  That is less than 1/2 drop.  You would have trouble measuring out anything less.

Scott

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.