Advia 2120i


Yolandalau

Recommended Posts

Does anyone one use Advia 2120 can tell me if  [1] Manual open tube sampler, [2] manual closed tube sampler and [3] automated closed tube sampler (Autosampler) are using the same path after aspiration?? Do we have to do blood comparison (correlation) for different modes?

Thanks so so much!!!

Link to comment
Share on other sites

We haven't had the 2120 for 5 years but yes, unless something has changed with that analyzer, all three modes have separate pathways thus the comparison between modes is a requirement.  We also had to run controls on all three modes as well.  That analyzer was a bear.  I hope you have better luck out of yours than we did... very needy..!!

Link to comment
Share on other sites