MARINA Posted October 10, 2017 Share Posted October 10, 2017 Is there a requirement to perform tachometer checks on a cytospin? The CFR's only state centrifuges and serofuges. Is it specifically required by CAP? Link to comment Share on other sites More sharing options...
SMILLER Posted October 11, 2017 Share Posted October 11, 2017 A cytospin is a centrifuge. Scott nziegler and exlimey 2 Link to comment Share on other sites More sharing options...
carol1 Posted October 17, 2017 Share Posted October 17, 2017 As SMILLER noted, a cytospin is a centrifuge. Here is the CAP requirement from the 08-21-2017 Lab General checklist: GEN.41017 Centrifuge Operating Speeds Phase II The operating speeds of centrifuges are checked at least annually as needed for the intended use, and this is done in a safe manner. NOTE: For centrifuges with safety mechanism preventing the opening of the lid while in operation, the checks of rpm should be performed only by an authorized service representative of the manufacturer or an appropriately trained clinical engineer. Evidence of Compliance: ✓ Records of verification of operating speeds at least annually REFERENCES 1) Clinical and Laboratory Standards Institute. Laboratory Instrument Implementation, Verification, and Maintenance; Approved Guideline. CLSI Document GP31-A. Clinical and Laboratory Standards Institute, Wayne, PA; 2009. Link to comment Share on other sites More sharing options...