As SMILLER noted, a cytospin is a centrifuge.  Here is the CAP requirement from the 08-21-2017 Lab General checklist:

GEN.41017

Centrifuge Operating Speeds

Phase II

 

The operating speeds of centrifuges are checked at least annually as needed for the intended use, and this is done in a safe manner.

NOTE:  For centrifuges with safety mechanism preventing the opening of the lid while in operation, the checks of rpm should be performed only by an authorized service representative of the manufacturer or an appropriately trained clinical engineer.

Evidence of Compliance:

      Records of verification of operating speeds at least annually

 

REFERENCES

1)

Clinical and Laboratory Standards Institute. Laboratory Instrument Implementation, Verification, and Maintenance; Approved Guideline. CLSI Document GP31-A. Clinical and Laboratory Standards Institute, Wayne, PA; 2009.

Link to comment