Daily Reagent QC requirements


CMCDCHI

Recommended Posts

CAP states:  "There are records of acceptable reactivity and specificity of typing sera and reagent cells on each day of use, including a check against known positive and negative cells or antisera, or manufacturer's instructions for daily quality control are followed."  Immucor package inserts for several reagents do not include instructions for a negative control.  Do we follow what Immucor suggests or add in a negative control too?  We currently test antisera against eachother for our daily QC, but are considering changing to Immucor corQC, but noticed there is no negative control for some reagents.  For those of you using corQC, how do you address this?  thanks!

Link to comment
Share on other sites

We keep 3% screening cells as well as our 0.8% screening cells. Most of the time our 3% screening cells suffice; however, on occasion, we find a cell negative for the anti-sera we're using on one of our panels or expired panels. Hope that helps!

Link to comment
Share on other sites

Just be wary -- a particular manufacturer's insert may not state that you need to run a particular control, but if your regulator requires one, you better have a better excuse than "they do not tell us to do it".  Inspectors always consider manufacturer's directions as minimum requirements. 

Scott

Link to comment
Share on other sites

We were cited by CMS for not having negative control for Anti A, Anti B, Referencecells and Selectogens 1-3.

We had just had our AABB/CAP inspection with no issues and felt we followed Immucor package insert. 

I would add negative QC because it is simple and I could not win the battle

Link to comment
Share on other sites

  • 4 weeks later...

I switched us to the Quotient QC kit for tube testing. It gives you a negative control for anti-A, -B, -A,B and -D. I freeze aliquots of AB plasma from patients for a negative control for A and B cells. My negative antibody screen control is a sample run the previous day on the Echo that had a negative screen - and it also serves as part of method comparison for antibody screens.

Link to comment
Share on other sites

Oh dear, whatever happened to common sense?

If you are testing  an antibody screen, for example, why do you need a negative control?  The majority of your samples will give a negative result.  So you know that the negatives are working.  But you DO need a positive control to make sure  that the reagents are working.  similarly, if you have an anti-k reagent, where are you going to get a k-neg cell to use as a control each time (in a normal lab)?  For ABO, if you have an A, a B and an O you have covered everything, even the reverse group, as with the A and the B one of the reverse group cells will be negative each time.....

Link to comment
Share on other sites

Really, then you don't run an auto-control with new AB Pos patients becayse A, B and O cover "everything".

Its going to come down to what your regulator pulls up regarding whatever standards they use: FDA, CAP, etc.  Granted, some of them do not make sense when you think of "science" stuff like serology and phyisiology, but you have to pass those inspections! 

Has anyone ever talked an inspector out of something that they have a standard for?

Scott

Link to comment
Share on other sites

if your laboratory is CLIA licensed, as most labs in the US are, you are required to follow the CLIA regulations.

The CFR requires a negative control. It is found in the CLIA regulations 42 CFR 493.1256 Standard: Control Procedures. 

42 CFR 493.1256 (c)(2) For each test system, perform control procedures using the number and frequency specified by the manufacturer or established by the laboratory when they meet or exceed the requirements in paragraph (d)(3) of this section.

(d)(3) At least once each day patient specimens are assayed or examined perform the following for -
(i) Each quantitative procedure, include two control materials of different concentrations;
(ii) Each qualitative procedure, include a negative and positive control material.

Further down in that section, there are requirements to test controls before resuming patient testing when a complete change of reagents is introduced... to rotate control material testing among all operators who perform the test, and to test control materials in the same manner as patient specimens.

Link to comment
Share on other sites

On ‎09‎/‎29‎/‎2017 at 12:21 PM, Kim D said:

We were cited by CMS for not having negative control for Anti A, Anti B, Referencecells and Selectogens 1-3.

We had just had our AABB/CAP inspection with no issues and felt we followed Immucor package insert. 

I would add negative QC because it is simple and I could not win the battle

Classically the Feds have only required positive controls with the exception of anti-D QC which also requires negative controls.  If you go into CFR used by CMS you probably will find that they require a pos and neg control for all moieties being tested.

Link to comment
Share on other sites

7 hours ago, galvania said:

Oh dear, whatever happened to common sense?

If you are testing  an antibody screen, for example, why do you need a negative control?  The majority of your samples will give a negative result.  So you know that the negatives are working.  But you DO need a positive control to make sure  that the reagents are working.  similarly, if you have an anti-k reagent, where are you going to get a k-neg cell to use as a control each time (in a normal lab)?  For ABO, if you have an A, a B and an O you have covered everything, even the reverse group, as with the A and the B one of the reverse group cells will be negative each time.....

Daily QC kits include antisera for a positive control. For purposes of method comparison I also run a specific number of positive antibody screens from the Echo by tube using all the testing methods we utilize (LISS, Peg, and saline AGT).

 

When I was a student, I won't say how long ago, we had a plexiglass rack that held in excess of 40 tubes that were used for blood bank QC. Then over time the number of tubes needed for QC dwindled down to a single handfull. Now we are headed back in the other direction. My QC currently requires 27 tubes, but I just passed an inspection, so it's all good...I guess.

Edited by AMcCord
Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.