How are people checking new blood bank reagent lots and shipments to satisfy CAP standard com.30450 "New reagent lots and shipments  are checked against old reagent lots or with suitable reference material before or concurrently with being placed in service."  What kinds of samples are you using?  Is anyone using patient samples or external QC materials?

Link to comment
Share on other sites

1 hour ago, SMILLER said:

I would say that for many BB applications, the bit about "or with suitable reference material" would include adequacy by using standard QC materials before testing patients.

Scott

Who defines what is suitable?  The person being inspected or the person doing the inspection?  Sadly, in my experience, some of those doing the inspecting, particularly if blood transfusion is not their primary discipline (or, in one famous case, had never worked in blood transfusion in his entire life) are not up to speed with what is or is not suitable, and, as far as their knowledge of blood transfusion is concerned, are not the "sharpest knife in the canteen of cutlery", but have all the power, and are determined to use it at all costs.

Link to comment
Share on other sites

Malcolm makes a good point -- at least so far as you really do need to check with your regulators (CAP, JCAHO, FDA) before committing yourself to any P&P.  I have found that CAP inspectors are knowledgeable (they use acting lab people for inspectors) and do indeed like to work with labs on questions like these, and are not necessarily keen on proving anything.

Anyway, check your standards, be they CAP or whatever, they should have examples of compliance with each one.

Scott

Link to comment
Share on other sites

All reagents are QC'd in accordance with the package inserts. We use Immucor reagents so our we use "corQC" to QC traditional reagents (ABO, Rh, antibody screen). We use reagent screen or panel cells for Pos/Neg QC for antigen typing sera. The only lot-to-lot QC we do is for Fetal Bleed Screen Kit with a lot-to-lot QC. We haven't had any problems with CAP inspectors with this QC plan.

Link to comment
Share on other sites

Thanks for your replies.  I posed this question to CAP and their reply was in line with your response above in that this standard (com.30450) is relevant to the fetal blood screen kits which contain controls.   Do you have an external control material that you are using to compare lots for the FMH kits?   We have thought of using patient specimens, but we rarely encounter a positive patient.

Link to comment
Share on other sites

When we receive new reagents, we crosscheck with the current anti-sera and use cells currently in use as well as the cells that most recently expired.

Crosschecking the fetal bleed screen kits (we use FMH RapidScreen - Immucor), we use the new anti-D with the old kit cells (controls and indicator cells) and the old anti-D with the new kit cells.

We crosscheck the antibody panels when they are received against the panel currently in use using diluted anti-D reagent to give 1+ reactions.

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.