We are looking at the Ortho Vision. However, it appears the Ortho QC is somewhat expensive at around $500 a month. At least here in Canada.

Has anyone had success with a home made recipe? We make our own QC in house for the Provue but am told it won't work with the Vision?

Appreciate your pros and cons on the instrument as well if anyone has a minute.

Thanks,

 

Richard

 

Link to comment
Share on other sites

  • 1 month later...
  • 9 months later...

You can now edit QC on the Vision.  You may also use User Defined QC.  Whole blood samples for Blood Bank and hemolysis can be problematic with any vendor or homemade creation.  The AlbaQ has less problem when stored in the upright position during storage, even the vials not yet in use.  It is very reliable and there's the ease of use with standard results and barcoding.

Link to comment
Share on other sites

  • 7 months later...

We made our own QC for Provue and had very minimal problems. I have continued to made my own QC for Vision with no problems as long as I am picky about the quality of plasma I pool and use. Can't have any hemolysis or lipemia. I do have the ability to edit QC results on the Vision and this is helpful of course. I know our rep and Ortho in general were very very pushy about buying their QC, saying it wouldn't work and "nobody does that". I beg to differ since about half of my training class DID in fact make their own. It is a huge savings to make your own for sure!

Link to comment
Share on other sites

  • 1 month later...
  • 3 weeks later...

We just went live with our Vision and are purchasing the AlbaQ - it is expensive and I wasn't happy to learn the open expiry is only 7 days!

Will definitely consider making our own once we are settled in to the new instrument...now that I know this has worked well for others.

Link to comment
Share on other sites

  • 5 months later...

Have 2 blood types that give a pos & neg reaction in each gel column, like the A neg and B pos of Albaq accomplishes.  For a 3 cell screen, an Anti D in a tube and an anti c in a tube.  Use plasma or a plasma like matrix, not just saline.  The number of drops of antisera depends on the strength or reactivity you want in your screens.

 

Link to comment
Share on other sites

10 hours ago, ALLOURBESTNLR said:

Have 2 blood types that give a pos & neg reaction in each gel column, like the A neg and B pos of Albaq accomplishes.  For a 3 cell screen, an Anti D in a tube and an anti c in a tube.  Use plasma or a plasma like matrix, not just saline.  The number of drops of antisera depends on the strength or reactivity you want in your screens.

 

I would use plasma from patients that have produced those antibodies instead of antisera.   The commercial antisera is usually not IgG.  

Link to comment
Share on other sites

1 hour ago, Kerry said:

R1R2 thank you! Do you have a formal policy and procedure that I can adapt and present to our medical director? I am a BB specialist so I want to make sure I’m doing this correctly.

I am not a BBer and  don't have a policy to share but I am sure others here can share. 

Link to comment
Share on other sites

  • 9 months later...

we prepare our own QC for the vision, and it's working perfectly.. because sometimes the ortho confidence is not coming regularly and delay in the shipment occur.

our QC goes like this :

your will need a :

- patient sample in EDTA with blood group of O  and call it QC1

- patient sample in EDTA with blood group of B and call it QC2

- patient sample in EDTA with blood group of A and call it QC3

- patient sample in EDTA with blood group of AB and call it QC4

Regradless of the Rh, but one of them must be Rh negative and Du must be performed for detection.

----------------------------------------------------------

for ICT : the rule is to check every Panel 1 to 3 if they are working properly :

so it goes like this :

- place an anti-D in plan tube and mark it as QC5 = mostly will give you panel 1 and 2 ,,, so we need one addition antibody to test panel 3

- place 5ml a different antibody in plan tube as long as it's testing panel 3  this will complete the reaction of the ICT panel 1.2.3 and mark this and call it QC6

---------------------------------------------------------

For phenotyping

 - use a patient sample and use it as a Rh phenotype sample EDTA and if it could be K+ ag that would be better call it QC7

---------------------------------------------------------

 

 

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.