Issues with Immucor Echo Screen Strips


Candybar

Recommended Posts

We are on Lot# R632 - no problems here, but we have seen this cell #2 problem in the past.  Changed out everything with no real resolution until we got a new Lot# of WBcorQC, since in our case, it was mostly the QC that was failing, patients were better.  If you are also seeing patient failures,  my best recommendation would be to get Service out to look the instrument over well (mini-tuneup) and get the newest lot# of RS3 strips Immucor can give you.   Best wishes.

Link to comment
Share on other sites

Our QC has been fine. The patients have been the problem. And the majority (70% ish) of the problem patients are prenatals. Our backup is gel which has been negative for all of them. But in the last few months we've had 2 weak (1-2+)  positive Echo screens which were negative on gel and turned out to be anti-E per our reference lab.

Immucor is sending us a new lot of indicator cells to try with both of our RS3 lots.

Link to comment
Share on other sites

I ran 9 prenatals which were negative with R643 using R630 strips. 2 were neg, 2 were pos and 5 were No Int. Most of our problems seem to be with pregnant women. Immucor is replacing them for us, pending approval.

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.