Freezer out of Temp


Recommended Posts

I came in this morning and noticed a huge spike on our -30 freezer temp chart from the evening prior.  The Freezer was at -12 for about 2 hours.  We have FFP and Cryo in there.  Does this mean I have to discard my entire inventory? Im having a hard time finding an answer via the technical manual

 

Any advice/input is appreciated!

Link to comment
Share on other sites

I wouldn't discard out of hand. We had a similar situation where we were temporarily storing frozen products in a freezer that did not have continuous monitoring and a tech reported a temp of -17C. Our supervisor submitted a biologic deviation report to CBER/FDA and we were able to use the products, which were obviously still frozen.

 

I would quarantine everything and check into this with the FDA -- while your products were a bit warmer than ours I would say that -12C should still be pretty frozen. :rolleyes: I believe that we quarantined until the report was reviewed and the products were released. Seems a shame to toss everything without looking into it.

Link to comment
Share on other sites

Thanks all for your input.

According to my blood supplier, I should be ok with a visual inspection of the units so long as they didnt look like they thawed.  Nothing was said of a FDA Biologically deviation report so I will def be looking into that. 

Link to comment
Share on other sites

Here's a quick link to the FDA page regarding it with the guidelines. It includes this verbage in section B.

 

"What Do I Report? (21 CFR 606.171(B))

  1. Under 21 CFR 606.171(B), you must report any event and any information relevant to the event associated with the manufacturing, to include testing, processing, packing, labeling, or storage, or with the holding or distribution, of both licensed and unlicensed blood or blood components, including Source Plasma, if that event meets all the following criteria:
  2. Either
    1. Represents a deviation from current good manufacturing practice, applicable regulations, applicable standards, or established specifications that may affect the safety, purity, or potency of that product; or
    2. Represents an unexpected or unforeseeable event that may affect the safety, purity, or potency of that product; and
  3. Occurs in your facility or a facility under contract with you; and
  4. Involves a distributed blood or blood component.

An adequate procedure for deviation reporting (21 CFR 606 . 100(B)) would include steps for determining whether or not an event is one for which a report must be submitted. The decision to report should be based on whether the event had the potential to affect the safety, purity, or potency of a product. The terms safety, purity, and potency are defined in 21 CFR 600.3(p), ®, and (s), respectively."

 

Better safe than sorry! The most time consuming part for us was logging all the unit numbers from the freezer. :unsure:

Link to comment
Share on other sites

Sadly, I think you will have to discard all of the products. They were well outside the minimum temperature for quite a while. And there needs to be corrective action for the tech involved as well as preventative action so this never happens again. Hide the key maybe?

Link to comment
Share on other sites

Thanks for all the input.  As of now our inventory is Quarantined with a small stock of new frozen product for emergencies.  

I had sent a few suspicious looking products back to the supplier to look at, and they aggreed and put them through their review process.  So we are waiting to speak with them

I cant imagine that being at -10 for 2 hours would thaw the product that much, but I want to be sure.

Thankfully no one was transfused with any frozen product within that time period or after.

Link to comment
Share on other sites

Ahhh!  We just had this happen.  The freezer completely died and the alarm failed to go off.  By the time it was discovered, the products were all at -8 and had been out of temp for hours.  We discarded all of the products.  Turns out that engineering knocked a wire loose when working on the motor and our wireless alarm failed to be noticed by security because someone had closed out the program....Needless to say, I was NOT a happy camper. 

 

This also happened about  8 years ago when someone had turned off the alarm. No one fessed up to it.... When dayshift came in all the products were thawed and freezer felt more like an incubator.  At the time, the freezer was located in another department.  We since have gotten a new lab where we actually have room for all of our equipment!

Link to comment
Share on other sites

Ahhh!  We just had this happen.  The freezer completely died and the alarm failed to go off.  By the time it was discovered, the products were all at -8 and had been out of temp for hours.  We discarded all of the products.  Turns out that engineering knocked a wire loose when working on the motor and our wireless alarm failed to be noticed by security because someone had closed out the program....Needless to say, I was NOT a happy camper. 

 

This also happened about  8 years ago when someone had turned off the alarm. No one fessed up to it.... When dayshift came in all the products were thawed and freezer felt more like an incubator.  At the time, the freezer was located in another department.  We since have gotten a new lab where we actually have room for all of our equipment!

 

Wow. This is one of the reasons why our automated temperature monitoring system has an alarm point and computer access to monitoring directly in the blood bank. If an alarm goes off it starts playing Rock and Roll by Led Zeppelin (clicked all the way to 11).

Link to comment
Share on other sites

Wow. This is one of the reasons why our automated temperature monitoring system has an alarm point and computer access to monitoring directly in the blood bank. If an alarm goes off it starts playing Rock and Roll by Led Zeppelin (clicked all the way to 11).

Yeah, that would be wonderful.  The best ours can do is be set up so that a pop up box shows on the computer.  None of our computers are equipped with speakers (thanks hospital IT!) and in order to see it, the program has to be open and running on the exact computer you are using.  This is why it is supposed to alarm in security.  Email notifications also go out, but if you aren't looking at your email....anyway, it isnt the best system.  Hind sight is 20/20!

Link to comment
Share on other sites

  • 7 years later...

When temp increases to the threshold set point (usually 0.5’C off limit), everything gets moved to a different freezer or shipping containers with dry ice. If you get dry ice from your blood supplier your set point may need to be lower to give a transport buffer. If using containers the temp is monitored every four hours. In other words everything is moved before it goes out of temp so nothing is thrown out. 

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.