I posted this question on the AABB site, but haven't received a response as of yet. (Does it take a long time to get a response from them?) So, I shall turn to my colleagues!

How frequently are we to activate the alarm on the blood bank refrigerators and freezers? I'm doing it quarterly, but is that what the regs require? If we can do it less frequently, I will gladly revise my method!

Have a Safe 2012!

Link to comment
Share on other sites

I just checked the AABB manual, 17th edition, and it says alarm activation quarterly, which means physically raising and lowering the sensor probe temps. We do a battery backup activation check daily. I didn't know we had to check the high and low electronic alarms daily. Is that a CAP requirement?

Link to comment
Share on other sites

hi we check our fridge alarms once a week, we check them the same day we change the chart recordings.

We do this by either leaving the door open, we have a remote alarm connected and wait until this has registered before closing it. We were inspected by MHRA this year and they were fairly happy with the system

Link to comment
Share on other sites

I just checked the AABB manual, 17th edition, and it says alarm activation quarterly, which means physically raising and lowering the sensor probe temps. We do a battery backup activation check daily. I didn't know we had to check the high and low electronic alarms daily. Is that a CAP requirement?

There has been much debate about whether the "physical" testing of probe temps is necessary on the Helmer refrigerators. Theory is...once you do a full validation that the high and low activation work properly and you verify temps with a certified thermometer, it's not necessary to put the probe in ice water, etc. Some inspectors are OK with this, some are not.

What other new method (for example, electonic crossmatching) do we fully validate, and then keep "going back to the old way" to make sure it still works? You could certainly argue this from both points of view.

Link to comment
Share on other sites

I would think it would come down to whether the probe itself can change its sensitivity over time so would read a different temperature (this change would be caught at annual calibration). The next question would be whether the alarm system can change what temp. it alarms at over time (i.e. at what signal level (voltage?) sent by the probe does it respond). I would think that this is more of a risk from humans making changes to the settings than from the system itself changing over time, but I am no expert. I suspect those Labs that test daily or weekly are really trying to make sure no one has disabled the system rather than that it goes off at exactly the right temperature.

Link to comment
Share on other sites

  • 1 month later...

CLSI (used to be NCCLS) has a Tentative Guideline: Temperature Monitoring and Recording in Blood Banks. NCCLS Document 16-T Vol. 6 no. 19 hints that alarm sensors should be tested monthly. 5.3.1(2) Platelet Incubator-Rotator but then uses the term quarterly for blood refrigerator and freezers. I have requested clarification from CLSI, but have not heard back yet.

Link to comment
Share on other sites

We test weekly but more to check that the staff on our switchboard know what to do. We test the procedure rather than the alarm. If there are alerts directly then quarterly 'should' be fine... I do always worry though that the actual alarm will fail - paranoid Annie me ;)

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.