QC of saline cubes attached to cell washers


Julie Shannon

Recommended Posts

We are using S/P Cat # B3158-1 and there is a statement on the cube saying that a negative control must be run daily at Coombs.

One issue is that we have 12 washers and rarely would each be used in a calendar day. But - since we do not log the station but rather the QC rack, we can't prove that a station is NIU. We work in circular pods with lazy susans so we generally share reagent racks.

Second issue - what is a suitable negative control? I am thinking just a negative DAT should be sufficient? Maybe one drop of a reverse or screen cell?

I am stumped on exactly how to implement this so I figured I would ask the subject matter experts.

Link to comment
Share on other sites

Think you just need to run a sample that has inert plasma and red cells, and then, once washed, add your AHG, spin gently (should be negative) and then add IgG-coated red cells, mix, spin gently again, and should be positive.

It's just to ensure that the wash is efficient. Sadly, to do it properly, you should do it for each bay in each of your washers!

Link to comment
Share on other sites

We run a weak D negative control daily, ending with a tube coombs, so i would think we would be covered. Except I am pretty sure that we do not record our lot numbers of saline anyway. Probably should go read what it says on our saline cubes. Never been an issue with inspectors as far as I know.

Link to comment
Share on other sites

My saline simply says 'inspect visually on a daily basis' for discoloration and particulates. I have a box for a checkmark for daily visual inspection of the saline along with all other reagents in our rack. We also perform a negative AHG test post incubation/wash, so I think that would cover false positives due to bacterial contamination, if an inspector asked.

Link to comment
Share on other sites

Why not just verify "negativity" with your first patient run of the day . . . assuming the pt will have a negative absc (or whatever). It seems this is like proving my balance for donors is working correctly. I weigh the 1st unit collected each day. It has to fall into defined parameters - if it does, then my balance is working fine. Why make extra work when you are already in compliance - JUST DEFINE IT!

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.