Donor Untested Autologous units


Recommended Posts

Where does the "DONOR UNTESTED" phrase go on ISBT-128 units? We collect autologous units for in-house use only and do not perform any disease testing. I can find no way to add that phrase using HemaTrax. Is using the AUTOLOGOUS USE ONLY, BIOHAZARDOUS an option? It seems that should be reserved for use if the patient tests positive.

Link to comment
Share on other sites

I am curious as to why you need this statement. We also draw autologous for in-house use only and they are untested. We are inspected by the FDA and have never had an issue with not having this statement on our units. I agree that it is not available in HemaTrax.

Link to comment
Share on other sites

AABB Reference Standard 5.1.6A-Requirements for Labeling Blood and Components. Item 23 Phrase "Donor Untested", if applicable*. *Footnote states: Donor not tested for evidence of infectious diseases.

From CAP Transfusion Medicine checklist:

Autologous units that are reactive or positive for ANY infectious disease marker, including a serologic test for syphilis, must be labeled with a "BIOHAZARD" label in addition to the usual labeling. Units that are prepared on site and are not tested must be labeled “DONOR UNTESTED.”

Link to comment
Share on other sites

Thank you! This is why I work in LIS and NOT the Blood Bank (anymore). But I am setting up Hema-Trax and Digi-Trax so your comment caught my eye. If I find out anything from Digi-Trax I will post.

Thanks again,

Laura

Link to comment
Share on other sites

Most facilites that are interfaced with HemaTrax are using Intended Use "1" Autologous Use Only Biohazardous for Untested Auto units. Although it sounds misleading, wouldn't you consider any tissue or blood specimen biohazardous unless otherwise tested?

Link to comment
Share on other sites

Up until several years ago we did label our autologous units "Biohazardous". When the AABB/CAP standards added the "Untested Donor" requirement for untested autologous units, we switched. Any thoughts on using the BIOHAZARDOUS statement with HemaTrax...is it acceptable to me more stringent in labeling or do we have to follow the requirements exactly?

Link to comment
Share on other sites

Thanks Joy...I was able to add it as a Special Test comment.

You can add it to the SpcTst.txt file in Program Files>Hematrax Standalone. Follow the format for what is already in that file (probably CMV Neg and CMV Pos) and give it next number in sequence. Save the updated SpcTst.txt file. This is what mine looked like:

CMV Negative=N0008,Negative for antibodies to CMV

CMV Positive=N0009,Positive for antibodies to CMV

DONOR UNTESTED=N0010,DONOR UNTESTED

After you do this, log back in to Hemtatrax and it should be in the Special Testing dropdown box. I have the standalone version. If you are interfaced with your BB computer system, please consult with Digi-Trax to make sure this change will not cause a problem.

Link to comment
Share on other sites

The problem I am seeing with doing this is that I am getting a barcode - that I cannot read of course. But those Special Testing "N" numbers are the designation from ICCBBA for the Special Testing General Database. So N0010 already has a defined Special Testing assigned comment (something about HLA antibodies). Not really sure if this matters since we can't send our untested autos outside of the institution but I wanted to make others aware.

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.