We haven't been having problems with our kits, but investigations for two survey failures for false positive results a few years ago pointed to the wash step as critical. Maybe start there. What works for us is a 12 x 75 tube filled to near the top, decant well after each of the 4 washes, make sure to break up the cell button prior to adding saline, and make sure that the saline addition is well mixed (uniform color throughout the tube). We had problems when we were using NERL saline. We now use unbuffered saline with pHix added (same that runs on the Echo). I had techs that tended to fill the tube only 2/3 full with saline for the washes and who weren't making sure the saline/cells were well mixed. Changing that habit seems to have helped. 

Link to comment
Share on other sites

Strange , we have never had that problem. We use immucor too.  I agree with AMcCord, washing is critical for this test.  Also, how are you counting the agglutinates?  By tube or slide? If you see agglutinates by tube, you should do it also on slide so you can count the number of agglutinates per LPF.   If you get less than 5 agglutinates in 5 fields, it is negative. 

Edited by SbbPerson
spellin
Link to comment
Share on other sites

We count under slide.  We wash with buffered saline, but not with PhiX added to unbuffered.  We purchase isotonic phosphate buffered saline from thermo scientific.  We have a sister hospital that follows the same procedure and reagents and they have not had issues.  Talking to Immucor we are the only customer that has reported issues with this.  We have gone through 3 lots.  I have checked the pH of the saline and it is 7.02, which is perfect.  We have washed both manually and with automated cell washers.  

Anyone have any additional thoughts or directions I should go?

Link to comment
Share on other sites

On 12/28/2022 at 8:33 AM, paddleking said:

We count under slide.  We wash with buffered saline, but not with PhiX added to unbuffered.  We purchase isotonic phosphate buffered saline from thermo scientific.  We have a sister hospital that follows the same procedure and reagents and they have not had issues.  Talking to Immucor we are the only customer that has reported issues with this.  We have gone through 3 lots.  I have checked the pH of the saline and it is 7.02, which is perfect.  We have washed both manually and with automated cell washers.  

Anyone have any additional thoughts or directions I should go?

Does your sister hospital uses the same Lot numbers of Immucor FBS kit? If yes, then you guys must be doing something wrong.  When was the last time your cell washer/centrifuge was calibrated? RPMs? Amount of saline dispensed? 

If they are not using the same Lot #'s, perhaps give them one of your kits and see if they get the same result. If the kit's QC passes for them, then there is something wrong with either your procedure or maybe equipment.  This is really strange. Good luck. 

 

 

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.