Body fluid validations


coppesml

Recommended Posts

We are looking to validate body fluids on our Coulter DxH 600. We don't get that many fluids - maybe 2 or 3 a week if we're lucky. I'm having trouble finding a product for validation studies. It seems as though we don't correlate very well manually :(. That is a huge reason for wanting to change to automated. What do you use?

Link to comment
Share on other sites

When we got our DxH 800s a few years ago, we tried to correlate several types of body fluids, and decided that it was not worth the trouble.    The linearity was not very precise (or accurate compared to manual counts) at lower levels, like those found in CSF.  You will have to do a manual diff anyway, as no hematology analyzer is going to report out mesothelial or malignant cells.

Scott

Link to comment
Share on other sites

  • 2 weeks later...

We too have a DxH600 it took a while but we did validate the instrument for body fluids.  The higher counts correlated better than the low counts.  Is there someone you can get samples from close by?  We basically manually count all clear samples as they do not correlate.  Our linearity is for body fluids is 10-66,000 for WBC and 150-5M for RBCs so CSF really cannot be done on the instrument unless it is cloudy.

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.