a,b positive in O baby with a Mum anti-D quant of 847IU


Kellimq

Recommended Posts

Hi Everyone, 

We did an IUT on a baby yesterday whose Mum has an anti-D quant of 847IU. Baby's pre-transfusion sample we tested on the Biovue card and the result was neg with the anti-A and anti-B but 4+ with the A,B. The D and the control well and DAT are also 4+. Any idea why the a,b is reacting with these cells? 

Thanks 

Kelli

 
Link to comment
Share on other sites

The DAT is easily explained, with an anti-D level that high!  Even after a double exchange transfusion, there will still be approximately 5% of foetal/neonatal red cells in the circulation, which is quite sufficient to give a positive DAT, but with an IUT, you are not removing an foetal red cells from the circulation (or minimal amounts) and should be considered more as a simple top-up transfusion (with profuse apologies to people who work in the Fetal Medicine Units, as I fully realise that there is nothing remotely "simple" about an IUT).

As far as the reactions with the anti-A,B, the anti-D and the control is concerned, while the anti-A and anti-B are negative, it must be remembered that there are potentiators in monoclonal antibodies, such as albumin, that will possibly lead to them giving "false positive" reactions (although, I would suggest, that the reaction with the anti-D is anything but false - unless the anti-D is blocking the D antigen sites, which, with an anti-D level of 847IUmL-1 [it is NOT 847IU, which is an amount, rather than a concentration] would not be in the realms of impossibility).  The amount of potentiator in each specificity will be different, depending upon what the antibodies are designed to detect (for example, what A and B subgroups, if any).  However, the anti-D will most certainly contain potentiators, to ensure that most weak, and some partial D types are detected.  This means that the control, which will have the same make up as the anti-D reagent - but without the anti-D, in other words, it would contain potentiators that would detect "false positives", which is the whole point of the control.

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.