You know, I hate to be the bearer of bad news but the manufacturer's instructions for gel cards do NOT support recentrifugation of the cards as a valid test method. There may be other reasons that results are not reproducible (e.g. fibrin, weak antibodies, variations in samples collected), but centrifuging the card again is not the way to resolve the problem! I hope no one is using this method to interpret patient test results!
From the insert: 1. False positive or false negative test results can occur from bacterial or chemical contamination of test materials, inadequate incubation time or temperature, improper centrifugation, improper storage of materials, or omission of test samples.
2. Proper centrifuge calibration is particularly important to the performance of the MTS Anti-IgG Card. TheMTS Centrifuge has been exclusively designed to provide the correct time, speed and angle.
and
For best results, it is recommended that following centrifugation, results should be read immediately. If tests are not read immediately, results may be affected by the drying out of the gel; hemolysis of the red cell and slanting of the reaction patterns due to storage in a non upright position.