CAP regulation? -mandatory audible alarm on blood bank refrigerators?


LIMPER55

Recommended Posts

Thanks, David.

 Our old refrigerator's audible alarms cannot be fixed.

We are going to rely on security to call us.

Here's the Joint Commission regulation; says the alarms must be audible. Time to get a new refrig. :)

QSA.05.04.03 : The laboratory uses alarm systems for refrigerators and freezers to monitor storage temperatures for blood and blood components.

Program: Laboratory

Chapter: Quality System Assessment for Nonwaived Testing

Rationale: N/A

Elements of Performance:

1. The laboratory has alarm systems for each refrigerator or freezer that meet the following requirements: - Alarms are audible. - Remote alarms are present for use when staff are not in the immediate area. - Alarms, including remote alarms, are monitored continuously. - The alarm system is battery operated or powered by a different circuit than the refrigerator(s) and freezer(s).

2. The laboratory has written policies and procedures for responding to the activation of the blood-storage alarm for refrigerators and freezers.

3. Policies and procedures for responding to the activation of the blood-storage alarm for refrigerators and freezers include the following: - A list of staff to notify, in order of priority - Backup or alternative provisions for blood storage - A process for maintaining records

4. The laboratory makes available to blood bank staff its policies and procedures for responding to the activation of the blood-storage alarm for refrigerators and freezers.

5. The laboratory follows its policies and procedures for responding to the activation of the blood-storage alarm for refrigerators and freezers.

Link to comment
Share on other sites

David,

When I read TRM.42750, it states "Is there an audible alarm for each component storage unit, is it continuously monitored 24 hours per day (in laboratory or remote) and has the response system to an alarm been validated?"

So, according to this, Yes, I would cite, unless there is a remote system in place, like Isensix, that has a responder to its audible alarm. Limper55 states that they rely on security to call them, so there must be a remote system in place.

Link to comment
Share on other sites

David,

When I read TRM.42750, it states "Is there an audible alarm for each component storage unit, is it continuously monitored 24 hours per day (in laboratory or remote) and has the response system to an alarm been validated?"

So, according to this, Yes, I would cite, unless there is a remote system in place, like Isensix, that has a responder to its audible alarm. Limper55 states that they rely on security to call them, so there must be a remote system in place.

Molly

quoting the July, 2015 changes to the standards (yes my inspection is soon):

 

"All component storage units are equipped with an alarm system that is monitored 24 hrs/day (in lab or remote), with alarm checks (for both low and high settings) performed and recorded at least quarterly.

Note:  the laboratory should be able to demonstrate how the alarm system works and that there is a process to ensure a timely response to an alarm, including remote alarms."

 

No mention of audible at all . . .  what version are you quoting from?   I don't do a low check on my freezer.

 

Like I tell the docs - I just report the data.

Edited by David Saikin
Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.